Auscultation is all audio. Following the bouncing needle will put you off into the ditch.

I absolutely get where you're coming from though. Intuitively it seems that the bounce of the needle should have some correspondence to the reading of the systolic b/p, and it probably does, just not in any way that is useful to you.

Good question.

Dwayne

Link to comment
Share on other sites

Dwayne is right. Do NOT rely on the bouncing needle- it is NOT accurate. If you want to prove that for yourself, do a little comparison. Auscultate a BP, palpate a BP, and then obtain one by watching the needle bounce on the gauge. Do this for multiple people, and compare the results.

Link to comment
Share on other sites

To check for accuracy of a unit we purchased and installed in the ambulance, we do similar comparisons. It's a unit that can do several things, not specifically made for EMS, cost as much as a nice used car, and stays put. However, we choose to only employ certain features. Also helps to weed out who can take an accurate BP, b/c it is possible to fake it for years - apparently - and be 'almost right' enough to get by. I did one of our 'self accuracy tests' after I read this earlier. Palp: 106/P; Scope: 110/62; Machine: 108/60.

Link to comment
Share on other sites

This thread is quite old. Please consider starting a new thread rather than reviving this one.

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

×
×
  • Create New...